Before you give up on the selenium cell itself, please be sure to check the meter. Many times I see these meters fail because of three factors not associated with a dead or dying receptor cell. In order: (1) the tapered ends of the rotating coil shaft are damaged (usually broken). The ends of the shaft are about as thin as a human hair, and when these meters/cameras are tossed on a bed, the back seat of a car, dropped, etc, enough g-force is generated to wreck them. If you see the shaft wobbling as you move it, the meter is gone. (2) the tiny wires leading from the rotating coil have broken. Unfortunately this is fatal, also. You may try to re-solder them, but it is about like I would expect microsurgery to be. (3) a loose piece of metal--could be a nut, small bolt, chip off of the magnetic core, anything, has become lodged between the magnet and the rotating coil. Especially in early German and Japanese meters, this is very common. Many of them were built as "open" meters...meaning no end covers. Anything that is attractive magnetically and floating around in the camera will go straight to the meter's magnet. Good luck. Some of these last a long time.
Jon