Hi I have a similar issue and I was wondering if any of you had figured out how to align the 28? Thanks!
As regards the 28 I went ahead and did as I indicated.
With 1.35V applied the light source was varied until there was a Full Scale (FS) deflection at 400 ASA. (600th sec)
I noted the deflection on the slowest ASA with the light at the same setting. Was between 60th and 30th. The light setting was left as it was for the rest of the work.
I then connected a variable resistor (22K) across the meter and brought the FS deflection back to 600th second with 1.55V applied. It was not correct at the bottom end of the scale.
😕
It took 300 ohms in series with the meter to bring the lowest ASA setting back to between 60 and 30th. This had almost nil effect on the FS ASA 400 setting. I suspect the adjustment next the meter is for this - it has a value of 1.5K ohms and would be ideal for just such fine tuning at the slower speeds.
I didn't touch it on mine but added the separate resistor in series alongside the shunt. Everything works fine now on a silver cell 1.55V.
To replace a non-functioning CdS cell unfortunately you'd need some sort of 'standard' light to set the 400 ASA/ FSD of the meter which is presumably what they have in the factory. I can't help there, sorry. With that they'd adjust the 400 FS deflection either with a ND filter (? exposed film?) or black blob on the CdS window and set the bottom end with the internal variable next the meter.
For more shunt/series resistor detail also see
http://www.kyphoto.com/classics/siloxidemodify.html where it is explained with diagrams how the values are derived mathematically - I did it in a more practical way, as in the first diagram of the 2 using a single variable resistor, found by trial and error, and measured the two 'arms' of the pot with a multimeter when I had both ends of the meter scale correct. Fixed resistors of the correct value were then substituted permanently. Getting small resistors is the hardest part! Surface mount would've been nice and small.
Murray